What happens to the nominal focal spot size as the x-ray tube target (anode) decreases?

Explanation:
The line-focus principle governs this. The size we see as the focal spot on the image (the nominal or effective focal spot) is a projection of the actual focal spot on the beveled anode. When the anode angle is decreased (more slanted toward the beam), that projection becomes longer across the beam path, so the apparent focal spot size increases. So, as the x-ray tube target angle decreases, the nominal focal spot size increases. This trades off resolution for better heat distribution: larger apparent spot means less sharpness but more even heat spreading.

The line-focus principle governs this. The size we see as the focal spot on the image (the nominal or effective focal spot) is a projection of the actual focal spot on the beveled anode. When the anode angle is decreased (more slanted toward the beam), that projection becomes longer across the beam path, so the apparent focal spot size increases. So, as the x-ray tube target angle decreases, the nominal focal spot size increases. This trades off resolution for better heat distribution: larger apparent spot means less sharpness but more even heat spreading.